Guys and Girls,
I know we have to bump test gas monitors before each use but I am struggling to find UK guidance and or legislation on this one.
Anyone looked at this recently and can point me in the right direction?
Thanks in advance.
|
|
|
|
Rank: Forum user
|
Below is an extract from HSE guidance on gas detectors (flammables)
42 The frequency of the function checks will depend on the manufacturer’s recommendations. For portable gas detectors, these should include checking:
■ that a zero reading is obtained in a clean atmosphere away from known sources of detectable gas or vapour;
■ for a response to a known flammable vapour;
■ the battery charge level is adequate for the job;
■ the level of any other reservoirs (eg hydrogen in a flame ionisation detector) is
adequate;
■ the logging period (if datalogging is required) to ensure all datapoints over the
required duration of measurement can be stored in memory.
Phil
|
|
|
|
Rank: New forum user
|
Thanks Phil, I saw this yesterday before posting on the forum, I am sure there has been a further note put out with the requirement to bump test daily. I will get in touch with our gas monitor supplier on Monday, they should have the data I am looking for. Thanks.
|
|
|
|
Rank: Forum user
|
check the manufacuturers documentation provided with the gas monitor, you should find the required information there (often in small print).
|
|
|
|