You don’t need to use sterile wipes to test water.
You can get calibrated prob thermometers for 20 pounds don’t bother with IR as expensive and break.
Here is the Techsupp I provide.
1. Introduction
This technical supplement has been written to ensure the Thermometers calibration is still within working limits.
1.1 Requirements
Temperature measurement and control devices must periodically be calibrated in order to prove their accuracy over the full operating temperature range and according to the device’s data sheet definitions. Proven accuracy is mandatory because inaccurate readings can lead to serious risks. Some devices are covered by calibration certificates from the device manufacturer. These certificates are valid for a defined period of time and the associated devices may be used throughout this period without additional calibration.
However, there are circumstances under which proper device calibration or re-calibration
is needed. The list below is not comprehensive but illustrates some of these
circumstances:
• A calibration certificate is not available because it has been lost.
• The device is used for longer than the period covered by the calibration certificate,
either at the user’s risk or with the approval of the manufacturer.
• The device was used or treated beyond the manufacturer's data sheet limitations
(e.g. excessive temperature, shock, etc.).
• The battery powering the device was replaced.
• The device's measurements are suspect.
• The device manufacturer specifies that a calibration procedure should be carried
out at regular intervals.
1.2 Objectives
Ideally, calibration should be carried out by a nationally or internationally accredited
calibration laboratory with proven accuracy standards and appropriate equipment.
Wherever possible, calibration should be carried out in accordance with the device
manufacturer’s instructions, or by following a device-specific policy.
2. Procedure: The ice-water bath provides an accurate reference temperature at 0°C if the melting ice -
water mixture is properly set up, handled and maintained. An accurate temperature is
achieved by this method because an ice-water mixture in a container which is open to the
atmosphere will stabilize at its own “triple point”. At this point all three aggregate states of
water coexist: liquid, solid and gaseous.
2.1 Action:
Place ice into a small clean jug or container and add a small amount of water to just cover the ice. Ideally chilled water should be used, the water level should be enough to saturate the ice but not enough to float it. Place thermometer probe into the slush for at least one minute, taking care not to let the probe contact the container. If the thermometer does not read between -0.5°C and +0.5°C of the reference thermometer, adjust it to 0°C. Non-adjustable thermometers outside of the + or – 2 degrees will be removed from use until they have been professionally serviced. Thermometers that have required adjustment for 3 consecutive months should be replaced.
2.2 Calibration in hot water:
Heat a clean container of water to a rolling boil. Place the thermometer probe into the boiling water for at least one minute, taking care not to let the probe contact the container. If the thermometer does not read between -0.5°C and +0.5°C of 100⁰C, adjust it to 100°C. Non-adjustable thermometers outside of the + or – 2 degrees will be removed from use until they have been professionally serviced. Thermometers that have required adjustment for 3 consecutive months should be replaced.
3.0 Record:
Record all findings on the Thermometer Calibration record.